    Analyzing Be (in HNO3 and HF) by GFAAS

    Tara Sheehan

      Currently our lab owns an AA800 Graphite Furnace.  We anlayze various trace metals by EPA Method 29, which requires a filter digestion using HNO3 and HF.  We have found that when we use the recommended conditions for the different steps in the analysis program, we receive a 0% spike recovery and a lower than usual concentration on a known concentration.

       

      Is there anyone out there using an AA800 and have had difficulty analyzing Be in a HNO3/HF solution?  Not sure what the drying, pyrolysis, and atomization should be in the method.  The recommended conditions do not work.

       

      I currently have:

       

      Temp:          Ramp:          Hold:

      100               5                    50

      120               30                  10

      1500             1                    5

      2400             0                    5

      2500             1                    5

       

      Sample at 10µl,  Diluent at 14µl

      Matrix modifier at 5µl Mg(NO3)2

       

      Thank you.